Understanding the principles and applications of spectrophotometry in biochemical analysis, including absorbance measurements, Beer-Lambert Law, and determining concentration of biomolecules.
Spectrophotometry – Quick Reference Notes
[ A = \varepsilon , l , c ]
| Component | Function |
|---|---|
| Light Source | Provides continuous radiation (e.g., deuterium lamp for UV, tungsten‑halogen for visible). |
| Monochromator / Filter | Selects a narrow wavelength band (λ) from the source. |
| Sample Holder (Cuvette) | Holds the solution; standard path length = 1 cm (square‑wave quartz or glass). |
| Detector | Converts transmitted light into an electrical signal (photodiode, PMT, CCD). |
| Readout/Display | Shows absorbance (A) or transmittance (T = I/I₀). Often linked to a computer for data storage. |
| Step | Action | Tips |
|---|---|---|
| 1. Calibration / Blank | Run a blank (solvent only) to set 0 % absorbance. | Use the same cuvette type as for samples. |
| 2. Wavelength Selection | Choose λₘₐₓ (maximum absorbance) from a spectrum scan. | Maximize sensitivity and minimize interference. |
| 3. Sample Preparation | Dilute sample so that expected A is between 0.1–0.8. | Avoid bubbles; use matched cuvettes for reference and sample. |
| 4. Measurement | Record absorbance at λₘₐₓ. | Perform duplicate/triplicate readings for precision. |
| 5. Quantitation | Use calibration curve (A vs. known c) or directly apply Beer‑Lambert law. | Verify linearity; include standards that bracket sample concentration. |
| Field | Example Uses |
|---|---|
| Biochemistry | Protein concentration (Bradford, BCA assays), nucleic acid quantification (260 nm). |
| Environmental Chemistry | Determination of nitrate, phosphate, heavy‑metal complexes. |